Syed Faiyaz
XG ERROR4CODE60 319
May 17, 2021 11:23
First XG Error phase is 4code 60-317H and
XG Error phase is4 code60.60- 319 H kv unbalanced is comming
I need help
to rate and post a comment
Maurice Dilick
re: XG ERROR4CODE60 319
May 18, 2021 09:58
Your HV Tank is bad. Hundreds available here.
to rate and post a comment
Carlos Duran
re: XG ERROR4CODE60 319
May 18, 2021 09:58
The Kv total comes from add anode with catode voltajes; they must be symmetrical. The unbalance means either is bigger than other.