ADEMETAN OLUWASEGUN

x axis controller loop error

November 09, 2012 10:08

hello colleagues, can anyone help out? my siemens magnetom c MRI during a scanning procedure of the neck suddenly developed some technical problems.The error message displayed was GRADIENT POWER AMPLIFIER ERROR. X-AXIS CONTROLLER LOOP ERROR OR CONTROLLER ERROR SIGNAL TOO HIGH.POWER STAGE'S TEMPERATURE TOO HIGH ( HEAT SINK OR AIR )
segun_ademetan@yahoo.co.uk
thanks

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: x axis controller loop error

November 09, 2012 05:33

Check your secondary chiller and connectors to the gradient check your chiller flow to the gradient.

Log inor Register

to rate and post a comment

(268)
(20)
(167)

Marshall Shannon

re: x axis controller loop error

November 10, 2012 12:13

Change your x gradient

1 person liked this
Log inor Register

to rate and post a comment

(60)
(2)

ADEMETAN OLUWASEGUN

re: re: x axis controller loop error

November 10, 2012 07:30

thanks !! what are the steps in doing that

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Philips Achieva 1.5T MRI Spare Parts Available Looking for Philips Achieva 1.5T MRI Parts?

We offer reliable & cost-effective spare parts for the Philips Achieva 1.5T MRI system helping you minimize downtime & keep operations smooth. Wide inventory available. Quick turnaround time. Expert support. Worldwide delivery. Contact Limor@elsmed.com

DOTmed Certified stats

Próximos eventos de leilão

Therapeutic Technologies
TIMED Ending Soon
Jun 15, 2026 (533 lots)
Cbrx2sales
TIMED
Jun 17, 2026 (127 lots)
Primis Medical
TIMED
Jun 18, 2026 (82 lots)
ZRG Medical
TIMED
Jun 26, 2026 (117 lots)