Manuel Casanova
SYTEC 4000i ERROR WARM UP
June 02, 2010 12:21
HI ALL:
I have a problem with SYTEC 4000i in warm up error:
err: 152 WARMUP ERROR LCP.03 TN WARM ADD. 002D3072
HPF2059
Thank´s in advance
to rate and post a comment
Elliot Marcus
Re: SYTEC 4000i ERROR WARM UP
June 02, 2010 01:23
Go to your support maintenance program and perform the Warm-Up Error Analysis. The HPF (the High Pass Filter) test is failing.
to rate and post a comment
Ronald Liem
Re: SYTEC 4000i ERROR WARM UP
June 06, 2010 10:05
Manuel,
Check the DAS PS +5V, +15V, -15V located below The DAS/Detector.
email me ronaldliem@yahoo.com