PROBLEM: VERTICAL THICK BANDS IN TOPOGRAM
-THICK CONCENTRIC RINGS IN TOMGRAM IMAGE.
-ASKING FOR FREQUENT RESET WHILE WARM UP,CALIBRATION,TOPOGRAM AND TOMOGRAM.
Previously got image artifact problems,but not frequent reset problem that time image artifacts gone by replacing the working DAB BOARD.
NOW THE ERROR MESSAGE
SMS IRS ERROR,BLOCK COUNT EXCEEDED,SCAN READING (1/502-1/502),SCANCOUNT 1 MAX COUNT 0
ANY SUGGESTIONS PLEASE.
Dear Mr. Srinivas
Thick line in top & ring in Tom. are typical problem from faulty or bad connection from Hybrid Pc (old name integrator poards).
In service mode by Ring test you can find the faulty faulty pc.
With Bests
H-Shirazi
emotion1333@yahoo.com
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Divi Srinivas
SOMATOM AR STAR PROBLEM
January 17, 2013 12:53
PROBLEM: VERTICAL THICK BANDS IN TOPOGRAM
-THICK CONCENTRIC RINGS IN TOMGRAM IMAGE.
-ASKING FOR FREQUENT RESET WHILE WARM UP,CALIBRATION,TOPOGRAM AND TOMOGRAM.
Previously got image artifact problems,but not frequent reset problem that time image artifacts gone by replacing the working DAB BOARD.
NOW THE ERROR MESSAGE
SMS IRS ERROR,BLOCK COUNT EXCEEDED,SCAN READING (1/502-1/502),SCANCOUNT 1 MAX COUNT 0
ANY SUGGESTIONS PLEASE.
to rate and post a comment
Hossein Shirazi
re: SOMATOM AR STAR PROBLEM
January 18, 2013 12:43
Dear Mr. Srinivas
Thick line in top & ring in Tom. are typical problem from faulty or bad connection from Hybrid Pc (old name integrator poards).
In service mode by Ring test you can find the faulty faulty pc.
With Bests
H-Shirazi
emotion1333@yahoo.com
to rate and post a comment
xiaohua xu
re: re: SOMATOM AR STAR PROBLEM
January 19, 2013 11:27
you are right
to rate and post a comment