SIEMENS SOMATOM AR STAR PROBLEM
June 13, 2012 09:44
i FOUND FOLLOWING PROBLEM WITH SOMATOM AR STRR--
1.WHILE TAKING TOPOGRAM VERTICAL BANDS ARE APPEARING OVER THE IMAGE
2.WHILE TAKING TOMOGRAM-CONCENTRIC THICK RINGS APPEARING
IMges are not interpretable .kindly anybody give suggestions for the problem
thanking you
to rate and post a comment
JORGE LUIS LOPEZ
re: SIEMENS SOMATOM AR STAR PROBLEM
June 14, 2012 12:47
If you already done air calibration, and the ring persist, it's possible that, that one o some of the cams of the detector are bad.
You can try doing air calibration first, and see if they disappear.
Hope this helps.
Jorge Lopez
to rate and post a comment
Juan de la Cruz Martinez
re: SIEMENS SOMATOM AR STAR PROBLEM
June 13, 2012 11:43
Have you try calibration? it seems like you might have bad detectors or a pcb on the system. process raw data to find out where in the system is the artifacts are coming from
to rate and post a comment
Hossein Shirazi
re: SIEMENS SOMATOM AR STAR PROBLEM
June 13, 2012 11:41
Dear Sir:
Cable betwwen D3 &Datector should be checked(Black cable with special connector), D3 itdelf can be faulty.test trasmit with sramble,increase & test pattern.
With Bests
H-Shirazi