Divi Srinivas

SIEMENS SOMATOM AR STAR PROBLEM

June 13, 2012 09:44

i FOUND FOLLOWING PROBLEM WITH SOMATOM AR STRR--
1.WHILE TAKING TOPOGRAM VERTICAL BANDS ARE APPEARING OVER THE IMAGE
2.WHILE TAKING TOMOGRAM-CONCENTRIC THICK RINGS APPEARING
IMges are not interpretable .kindly anybody give suggestions for the problem
thanking you

Log inor Register

to rate and post a comment

(100)
(9)
(4)

JORGE LUIS LOPEZ

re: SIEMENS SOMATOM AR STAR PROBLEM

June 14, 2012 12:47

If you already done air calibration, and the ring persist, it's possible that, that one o some of the cams of the detector are bad.
You can try doing air calibration first, and see if they disappear.
Hope this helps.
Jorge Lopez

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: SIEMENS SOMATOM AR STAR PROBLEM

June 13, 2012 11:43

Have you try calibration? it seems like you might have bad detectors or a pcb on the system. process raw data to find out where in the system is the artifacts are coming from

Log inor Register

to rate and post a comment

(7)

(1)

Hossein Shirazi

re: SIEMENS SOMATOM AR STAR PROBLEM

June 13, 2012 11:41

Dear Sir:
Cable betwwen D3 &Datector should be checked(Black cable with special connector), D3 itdelf can be faulty.test trasmit with sramble,increase & test pattern.
With Bests
H-Shirazi

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats

Próximos eventos de leilão

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)