Siemens Siremobil 2000 Error Code 500B
September 16, 2010 04:30
Does anyone have the error code listing for Siemens Siremobil 2000. Specifically looking for 500B but a full list would be helpful.
Thanks.
to rate and post a comment
alejandro mendoza
Re: Siemens Siremobil 2000 Error Code 500B
September 21, 2010 12:09
500B Overload current in intermediate-circuit reduct. contr
Cause : If the overcurrent monitor of intermediate-circuit
. reduction controller constantly responds during a
. preset period of time, an error is displayed
. (V1.1 IZ D202 lights up for approx. 0.5 s).
Active: during radiation, standby
Possible trouble / help : Short circuit in reduction
. controller bridge
. Malfunction of controller (ground wiring, shielding)
.
regards.....Alex.
to rate and post a comment
Mohmmad Samir Hawout
Re: Siemens Siremobil 2000 Error Code 500B
November 21, 2010 10:59
Hi, Sir.
Also I have Err : 2403 and Err : 0027 , please tell me , What I can doing for it ?