Standby Declaration: D916 detects actual kV error.
Action:
· Check the cable connection of measurement data log D907, the single tank, and D916.X8 (CPU).
Possible causes: Defective single tank, defective cable, defective D916 (CPU).
One call to 440-263-6575 and your move will be covered.
We repair & store trailers. We build to your custom needs.
JDI Mobile Medical Manufacturing. Inc. We have your back.
Ricardo Paulin
siemens mobilette error code 8
May 05, 2007 06:01
Does anybody knows what is the error code 8, in a Siemens Mobilette?
Thankyou
to rate and post a comment
Henrique Gomes
Re: siemens mobilette error code 8
May 07, 2007 09:21
If you have the Mobilett XP:
ERR 8 Static kV measured value 0
System mode:
Standby Declaration: D916 detects actual kV error.
Action:
· Check the cable connection of measurement data log D907, the single tank, and D916.X8 (CPU).
Possible causes: Defective single tank, defective cable, defective D916 (CPU).
Regards
Gomes.
to rate and post a comment
klaus boehmdorfer
re: Re: siemens mobilette error code 8
March 12, 2014 08:51
what type of Mobilett do you have?
to rate and post a comment