Siemens Mobilett Plus: Err 9

Tags: Radiology - R/F, SIEMENS

(10)
(1)

mdastech

Siemens Mobilett Plus: Err 9

January 20, 2016 01:41

Hi All,

We just installed a xray in a remote hospital. Whenever the X-ray is fired, it gives an Err 9. Does anyone know what might be causing this?

More info: The Xray machine is connected to a 5000 watts transformer to convert from 220 to 110V...could this be the problem?

Any help will be appreciated...thanks.

Log inor Register

to rate and post a comment

(10)
(1)

CARLOS FLORES

re: Siemens Mobilett Plus: Err 9

January 21, 2016 10:17


Hello The team works by charging capacitors, as have the 110 Volts AC supply there should be no problem.
  Information herror command you, you need to diagrams to work, I have a part but is incomplete.

I hope you serve and you solve your problem, my email to keep track of your progress and can be contacted as follows floresr_kco@hotmail.com

Err 9 During exposure When measuring the
kV value, a large high tension asymmetry
is detected.

Normally the kV value is symmetric around ground. The acceptable deviation is ± 19.5 kV. This corresponds
to HTD 2.5 V± 0.5 V.
· Intermittent measurement cables or defective single tank.
· Check the the measurement cable K9 between D1 and the single tank.
· Check the current to the single tank
· Measure the HTD signal on D1 and compare it with notes and curves.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Próximos eventos de leilão

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)