I have been facing err 21 on siemens mobilett plus. (The measured value of the standby filament current has
exceeded the limit.)
D4, +60 V, +15 V and +5 V are normal.
"FIL_VAL" on D1 must be between
0.16 and 2.5 V when the filament is
in standby mode. but mine is 12,5 volt.
Could you please assist me to solve the matter? Thank you
Before suspecting failure in H1, check cable integrity (short circuit) from D4 ( filament) to H1 (HV tank) ; open the filament circuit at Fil the connector on H1 , I sense must go to cero. Refer to wiring diagram document.
If not check D4 or cable broken
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
TechService
siemens mobilett plus err 21
February 12, 2025 03:39
Dear Friends,
I have been facing err 21 on siemens mobilett plus. (The measured value of the standby filament current has
exceeded the limit.)
D4, +60 V, +15 V and +5 V are normal.
"FIL_VAL" on D1 must be between
0.16 and 2.5 V when the filament is
in standby mode. but mine is 12,5 volt.
Could you please assist me to solve the matter? Thank you
to rate and post a comment
juan pablo amado
re: siemens mobilett plus err 21
February 18, 2025 10:28
Before suspecting failure in H1, check cable integrity (short circuit) from D4 ( filament) to H1 (HV tank) ; open the filament circuit at Fil the connector on H1 , I sense must go to cero. Refer to wiring diagram document.
If not check D4 or cable broken
to rate and post a comment
lei wang
re: siemens mobilett plus err 21
March 05, 2025 01:05
Remove the connection line between the filament plate and the combined ball tube, and turn on the computer to see if there is any change in the error.
to rate and post a comment
lei wang
re: siemens mobilett plus err 21
March 05, 2025 01:05
Remove the connection line between the filament plate and the combined ball tube, and turn on the computer to see if there is any change in the error.
to rate and post a comment