January 03, 2019 12:04

I am getting an Error on a Siemens SX 65/80. It is 715/080. Does anyone have experience with this error? Furthermore, I am looking for the software and a service manual. I heard the software will only on older Microsoft OS.

Log inor Register

to rate and post a comment

(18)

mahdi Bennai

re: Siemens Error 715/080

January 04, 2019 09:47

Hi. I can help
contact.medicotech@gmail.com

Log inor Register

to rate and post a comment

Lister

re: Siemens Error 715/080

January 04, 2019 09:47

Error 715 0
Meaning:
Minus-kVmin error
Description:
The minus kVmin monitoring becomes active if the -kV actual value is more than 10 kV
less than the -kV nominal value.
Display LED D100.V165 "kV MIN-"

Check both ends of the negative HT cable for signs of arcing, I have also had this problem in the past and the HT transformer itself was faulty.

1 person liked this