AHMEDELBAGA
seqouia 512
October 27, 2014 10:42
i have a problem with the system
it have a 6c2 transducer after a while when selecting the 6c2 probe system display error massage
"A hardware fault has occurred please call for service assistance"
after 2 days
when i power up the system with 6c2 connected it display the error massage any time i connect the probe the error massage displayed
i should remove it open system
i connect the probe to another seqouia machine it works well
any one can help????
to rate and post a comment
Rey Gonzalez
re: seqouia 512
October 27, 2014 06:30
do the other sequoia has the same software and pcb configuration? is it also a 512 or C512 share service? you may want to check the CN pcb and BF pcb rev in the system. Run diagnostics.
Hope it helps,
Rey
to rate and post a comment
AHMEDELBAGA
re: re: seqouia 512
October 28, 2014 09:47
the main machine C512 share service and probes was working well till this error ocured
the other one 512 GI where the probe works well