Hipolito Jorge Iglesias

senographe 600T error E60

August 22, 2009 02:05

dear Friends,
i will appreciate your help, i would like to know what error E60 means.
the machine does prep but no exposure
thank you

Log inor Register

to rate and post a comment

(174)
(8)
(31)

Hipolito Jorge Iglesias

Re: senographe 600T error E60

August 23, 2009 12:47

sorry, it is error E50

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: senographe 600T error E60

August 23, 2009 12:47

Hipolito:

Problem the communication between 400pl12 CPU board to 100pl6 console board, check the flat cable.

Best regards