My Ct is GE Lightspeed 8 slice (SGI computer)
Now every tim try scan.....shower Scan Hardware stop scan
Run "High Voltage Functional " is OK
But Run "Diagnostic Data Collection"
In "Static X-ray off" is pass
In "static x-ray on" will stop scan
error message is 184154
Host:STC
Function:XRAY GENERATION:FinalnExposure Command
Time out on Exposure Command to X-ray on.
HW Error:Possibile hardware problem
Timeout is:300 mS
run DDC static x-ray off is pass
static x-ray on ......stop scan
But We found in Flash download
Update .....can finish and ne error show
Querying....will show STC unknow,ETC unknow,OBC unknow
My friend talk to me ,try load old states,but no help
We try Rotor Control.......pass
Check fiber optics lines from console to DAS. Be careful, visible light pass test does not work.
You must use infrared test. TV remote control and old mobile phones with IR sense cameras work only. Two person Job.
The most probable optical cable fault is on the rotating part. Try stationary scan, gantry rotation power off and connect optical cable directly from Console to DAS.
run DDC static x-ray off is pass
static x-ray on ......stop scan
But We found in Flash download
Update .....can finish and ne error show
Querying....will show STC unknow,ETC unknow,OBC unknow
My friend talk to me ,try load old states,but no help
We try Rotor Control.......pass
Check all fuses and boards in HEMRC assembly...and perform rotor diagnostic tests....
Check the OBC-HEMRC Interface board
Check the KV & MA Boards in the OBC...If possible swap the Artesyn II boards between the STC and OBC (mind the jumper settings on each)
steven
Scan Hardware stop scan
October 20, 2017 10:02
My Ct is GE Lightspeed 8 slice (SGI computer)
Now every tim try scan.....shower Scan Hardware stop scan
Run "High Voltage Functional " is OK
But Run "Diagnostic Data Collection"
In "Static X-ray off" is pass
In "static x-ray on" will stop scan
error message is 184154
Host:STC
Function:XRAY GENERATION:FinalnExposure Command
Time out on Exposure Command to X-ray on.
HW Error:Possibile hardware problem
Timeout is:300 mS
Has any one can give me suggest.
thank
steven
to rate and post a comment
Tibor Ivanszky
re: Scan Hardware stop scan
October 21, 2017 10:18
There must be a set of error, look at the first. The error you described is a result of other error most probably.
tibor@ivanszky.com
to rate and post a comment
Gabriel Cool
re: Scan Hardware stop scan
October 23, 2017 09:09
It can be problem with tube
How mane mas is done ?
to rate and post a comment
Raul Elizalde
re: Scan Hardware stop scan
October 23, 2017 09:09
Hello
Could you make Again x Ray in DDC, then click on last page and take picture to last 5 pages and send me
Raúl
raul.elizalde.sti@gmail.com
to rate and post a comment
steven
re: re: Scan Hardware stop scan
October 24, 2017 09:41
run DDC static x-ray off is pass
static x-ray on ......stop scan
But We found in Flash download
Update .....can finish and ne error show
Querying....will show STC unknow,ETC unknow,OBC unknow
My friend talk to me ,try load old states,but no help
We try Rotor Control.......pass
thank for your message
steven
to rate and post a comment
CRAIG WAMBURA
re: Scan Hardware stop scan
October 23, 2017 09:09
contact me craig@telemed-technologies.com and we work on this together
to rate and post a comment
Tibor Ivanszky
re: Scan Hardware stop scan
October 23, 2017 06:06
Check fiber optics lines from console to DAS. Be careful, visible light pass test does not work.
You must use infrared test. TV remote control and old mobile phones with IR sense cameras work only. Two person Job.
The most probable optical cable fault is on the rotating part. Try stationary scan, gantry rotation power off and connect optical cable directly from Console to DAS.
to rate and post a comment
steven
re: Scan Hardware stop scan
October 24, 2017 09:41
run DDC static x-ray off is pass
static x-ray on ......stop scan
But We found in Flash download
Update .....can finish and ne error show
Querying....will show STC unknow,ETC unknow,OBC unknow
My friend talk to me ,try load old states,but no help
We try Rotor Control.......pass
to rate and post a comment
CRAIG WAMBURA
re: Scan Hardware stop scan
October 24, 2017 10:21
Check all fuses and boards in HEMRC assembly...and perform rotor diagnostic tests....
Check the OBC-HEMRC Interface board
Check the KV & MA Boards in the OBC...If possible swap the Artesyn II boards between the STC and OBC (mind the jumper settings on each)
to rate and post a comment
Georgi Gaydov
re: Scan Hardware stop scan
September 16, 2018 02:47
Hi Steven,
I have the same problem.
How you solve your problem.
Thanks
Georgi
to rate and post a comment
Georgi Gaydov
re: Scan Hardware stop scan
September 16, 2018 02:50
Hi Steven,
I have the same problem.
How you solve your problem.
Thanks
Georgi