Dear friend, first of all please check that your CR is adjusted on the floor or table top correctly. Open the front door and see the small circular water level.
If this is leveled correctly then the problem is with your optical unit
[Error 06550- Polygon rotation error rotation of the polygon was not stable when the push plate unit moved to the read position.]
The recommendation producer [Replace the optical unit.]
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
Regius Model 110
May 09, 2014 01:54
need a service manual for Konica Minolta Regius 110 CR.
My problem is: error 06550 call service
to rate and post a comment
Mohammad Ali Haidary
re: Regius Model 110
May 11, 2014 08:36
Dear friend, first of all please check that your CR is adjusted on the floor or table top correctly. Open the front door and see the small circular water level.
If this is leveled correctly then the problem is with your optical unit
to rate and post a comment
fortronik
re: Regius Model 110
May 11, 2014 08:56
[Error 06550- Polygon rotation error rotation of the polygon was not stable when the push plate unit moved to the read position.]
The recommendation producer [Replace the optical unit.]
to rate and post a comment