QDR 4500A
March 22, 2014 10:53
I have a DEXA machine model qdr 4500A. I am facing aproblem that durin daily
QC with phantom system shows an error "19.03.2014 14:27:44 ErrorId: IDS_SCN_CHECKRETRYDCAL ErrorNo: 9237
Exit, repeat the QC procedure and check the phantom position carefully. System
failed. Check that the phantom is the ONLY object on the table.
Test failed - mean exceeds limits. An error has occurred in a CSFStream.
Anyone can help me out to resolve this issue.
Regards
to rate and post a comment
Desmond Johnson
re: QDR 4500A
March 22, 2014 09:25
Try the alignment of the tube to the detector in section 4 of the manual. The CSFStream is the data from the detector. Check carefully the final value of the gain. If you don't get enough the tube or XRC is probably bad.
to rate and post a comment
MRAdmin
re: QDR 4500A
March 29, 2014 11:41
Hi Mahfooz,
before doing any alignment, check the X-ray tube, there me be a lot of oil in the lexan cup.
if you need any help contact me bm_mae@yahoo.com
to rate and post a comment
mahfooz khan
re: re: QDR 4500A
March 31, 2014 09:12
Thank you very much for your attention. Yes, the oil was leaking from lexan cup which sealed by tightening the related screws. The problem was resolved by running QRDriver calibration. We will also change the gaskits of the tank, later on.
to rate and post a comment
MRAdmin
re: re: re: QDR 4500A
March 31, 2014 04:35
It is great to hear that.
to rate and post a comment
Jordan Samichkov
re: re: re: QDR 4500A
July 24, 2019 11:58
Hello Khan,
Please inform about the sizes of the gaskits of tank for Hologic QDR Series, or send me an e-mail: j.samichkov@medicom2000.com
Jordan