Proteus Service Manual says take off the cover & open up the front of the collimator.
"Verify the setting of the DIP switch on the D1-Board. The DIP switch S2 on the D1-board must be set as follows:
S2.1 = ON
S2.2 = OFF
S2.3 = ON
S2.4 = OFF
S2.5 = OFF
S2.6 = ON
S2.7 = ON
S2.8 = ON
Start the collimator calibration mode to get the calibration data stored in the EEPROM.
Note: Without calibration data in the EEPROM of the board D1 the collimator startup may fail.
For collimator calibration carry out the following steps:
a) Verify the DIP switch S2 on D1 (see above). It’s important that switch S2.4 is set to OFF.
b) Close the lamellas in both levels (longitudinal and transversal) manually.
c) Connect the collimator to the power supply.
d) Set dip switch S2.4 to ON
e) The collimator will start a calibration run. The lamellas will fully open to the photoelectric barriers and then
close to the minimum opening.
The calibration data will be stored in the EEPROM.
f) Verify the setting of the DIP switch on the D1-Board again. The new settings are
S2.1 = ON
S2.2 = OFF
S2.3 = ON
S2.4 = ON
S2.5 = OFF
S2.6 = ON
S2.7 = ON
S2.8 = ON
g) Now the collimator is full working.
h) Disconnect the power supply before closing the collimator."
That is usually all that is needed. If the sizes displayed are not correct for some reason, I would check the cassette trays first and then do the "size display" calibration as described in the service manual.
Pete..is there any chance you have the Proteus rad room service manual available? We just took a room over from warranty and GE is not responding to my requests for a manual. If you can send it...send to elton.swope@nmhs.org
thanks
http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=XR
If you need other Info I can supply some specific stuff.
staate.hayward@sodexo.com
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
doug westlake
proteus collimator
August 06, 2012 11:50
i need to replace a collimator on a proteus rad room….how do you do the cals for the auto sizing?? any help would be great
to rate and post a comment
Pete Schliebner
re: proteus collimator
August 07, 2012 06:38
Proteus Service Manual says take off the cover & open up the front of the collimator.
"Verify the setting of the DIP switch on the D1-Board. The DIP switch S2 on the D1-board must be set as follows:
S2.1 = ON
S2.2 = OFF
S2.3 = ON
S2.4 = OFF
S2.5 = OFF
S2.6 = ON
S2.7 = ON
S2.8 = ON
Start the collimator calibration mode to get the calibration data stored in the EEPROM.
Note: Without calibration data in the EEPROM of the board D1 the collimator startup may fail.
For collimator calibration carry out the following steps:
a) Verify the DIP switch S2 on D1 (see above). It’s important that switch S2.4 is set to OFF.
b) Close the lamellas in both levels (longitudinal and transversal) manually.
c) Connect the collimator to the power supply.
d) Set dip switch S2.4 to ON
e) The collimator will start a calibration run. The lamellas will fully open to the photoelectric barriers and then
close to the minimum opening.
The calibration data will be stored in the EEPROM.
f) Verify the setting of the DIP switch on the D1-Board again. The new settings are
S2.1 = ON
S2.2 = OFF
S2.3 = ON
S2.4 = ON
S2.5 = OFF
S2.6 = ON
S2.7 = ON
S2.8 = ON
g) Now the collimator is full working.
h) Disconnect the power supply before closing the collimator."
That is usually all that is needed. If the sizes displayed are not correct for some reason, I would check the cassette trays first and then do the "size display" calibration as described in the service manual.
to rate and post a comment
Elton Swope
re: re: proteus collimator
August 08, 2012 08:34
Pete..is there any chance you have the Proteus rad room service manual available? We just took a room over from warranty and GE is not responding to my requests for a manual. If you can send it...send to elton.swope@nmhs.org
thanks
to rate and post a comment
Staate Hayward
re: re: re: proteus collimator
August 08, 2012 09:06
They are available on the GE doc search page.
http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=XR
If you need other Info I can supply some specific stuff.
staate.hayward@sodexo.com
to rate and post a comment