Hi guys,
I started having this following error code when I was getting helical scans, then I started to do just axial scans with no issue, until now, that I can not pass the SET point before Starting the exposure, because the systems aborts the scan.

ER-101 Scan Error 53 with Axial, Helical scans and also in warm ups.
Any help would be appreciated, thanks.

1 person liked this
Log inor Register

to rate and post a comment

CRAIG WAMBURA

re: ProSpeed SX

July 29, 2016 07:00

The 101 is a general scan error, when you look through the log, do you have any XG Errors? The XG Error will point us to the defective hardware.
Is there rotor noise?
Carry out HV Diagnostics.

You can contact me on skype: craig.wambura

1 person liked this
Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: ProSpeed SX

July 29, 2016 07:00

Er 101 code 53 : XG Ready Response Timeout;Check Optical Cable(E-Rack-----HV Inv); Connect opticat cable securely or replace it.

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: ProSpeed SX

July 29, 2016 10:50

101 53 , 54 intermittent in the morning slip-ring check data brush,and air conditioned quality.

check the brushes and slip rings.

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: ProSpeed SX

August 16, 2016 09:23

Hi guys, thanks for your reply, I found a 240-120 transformer damaged in the CTVRC module, somehow the connector to the CTVRC board was shorted and burned, probably 2 of the 4 cables got loose and had a shortcut. I will be replacing the motor and see what it does, I'll talk to you guys later to explain what

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: ProSpeed SX

October 02, 2016 02:51

Hi guys, so I replaced the CTVRC module, and still have the same ERROR

ER-101 Scan Error 00000053 LCP=03 TN=SEQC ADD.=00245E62

Also, when I checked the OGP board and everytime I hit the SET button to begin the gantry rotation, the D2A green led on the OGP board blinks, until I reset the board or the system, goes to normal. Do not know if this error has to do with the same ER-101, does anybody know something about it?

Thanks.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: ProSpeed SX

October 02, 2016 02:51

Sorry the D2A ERRS led on the OGP board, is ON and not blinking/flashing... the one flashing is the D3A LOOPS led (but this is normal).

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Próximos eventos de leilão

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)