FranciscoG

ProSpeed SX Error 117 XG Error

March 31, 2016 12:17

Hello, I have this error code in the ProSpeed

ER-117 XG Error 00000541, according to the diagnostic manual, there is an Anode Overcurrent and it wont make an exposure, does that mean the HV Tank is bad? How can I check this manually?

Could a bad tube send that type of error?

Log inor Register

to rate and post a comment

(11)
(3)

HONG SOON KANG

re: ProSpeed SX Error 117 XG Error

March 31, 2016 10:04

almost the cause is from bad X-Ray Tube. if you want to check in further, You had better execute no load test. but the best way in my opinion, is changing tube. if you need any help. just mail me. hskang@medihealth.or.kr

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: ProSpeed SX Error 117 XG Error

March 31, 2016 04:38

Hi Sir,

I took these measurments on the HV Cables:

X RAY TUBE Cathode side (-):
C-L : 0.5 ohm
C-S : 2.5 ohm
L-S : 2.5 ohm

X RAY TUBE Anode side (+):
C-L : 0.5 ohm
C-S : 0.7 ohm
L-S : 0.6 ohm

Anode HV TANK (+):
C-L : 0.3 ohm
C-S : 0.2 ohm
L-S : 0.2 ohm

Cathode HV TANK (-):
C-L : 0.2 ohm
C-S : 0.2 ohm
L-S : 0.2 ohm


Log inor Register

to rate and post a comment

(11)
(3)

HONG SOON KANG

re: re: re: ProSpeed SX Error 117 XG Error

April 11, 2016 08:49

The resistance between C-L, L-C, L-S should be 0

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: ProSpeed SX Error 117 XG Error

April 28, 2016 12:06

Hi I did change the Xray tube, checked the HV cables, had the HV tank (+) serviced cause it had air in it, and still can't make any exposures, and have the same XG error 541, besides this other error:

Scan Error 74
Scan Error 73

Sometimes is 74 another times 74.
Any help?

Log inor Register

to rate and post a comment

(5)

John Robert David

re: re: ProSpeed SX Error 117 XG Error

May 31, 2017 10:41

Hello sir. Can i ask how to perform no load test? Thanks

Log inor Register

to rate and post a comment

(123)
(12)
(7)

Manuel Brugat

re: ProSpeed SX Error 117 XG Error

April 03, 2016 01:10

Francisco:

Most likely your x-ray tube is gone. It probably is gassy. How old is this tube?

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: ProSpeed SX Error 117 XG Error

April 08, 2016 10:09

Hi Manuel,
The tube is a 2010 model, the rotor is very noisy as well, I took the +HV tank and it had air in it, so I sent it to service and still having the same error.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: ProSpeed SX Error 117 XG Error

April 13, 2016 06:04

I will be replacing the tube, thanks.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: ProSpeed SX Error 117 XG Error

July 09, 2016 04:32

So after changing the x ray tube, that was bad in first place, I ended up changing the Anode HV tank, and it got fixed. Thanks to everyone for your help.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: ProSpeed SX Error 117 XG Error

May 11, 2017 09:11

Hi again guys, I'm having a similar error again on this system, but with another error code:

ER-117 XG Error code 148
(Generator Safety Loop Open)

Does any one knows where the Generator Safety Loop is located?