Over 500 Total Lots Up For Auction at One Location - NJ 06/15

problem with ge voluson 730 pro

(182)
(31)

MIGUEL ANGEL TORRES MARTINEZ

problem with ge voluson 730 pro

June 17, 2016 03:02

When I connect the probe in the screen don't show image.

Log inor Register

to rate and post a comment

(21)
(2)

Pablo Delgado

re: problem with ge voluson 730 pro

June 17, 2016 06:06

Hi,
Did you try to test the probe in another system?
Did you get some kind of error?
The ultrasound system works with another probes?

Log inor Register

to rate and post a comment

(76)
(4)
(34)

Lakshmi Narayana

re: problem with ge voluson 730 pro

June 19, 2016 02:59

It happens with all the probes? If so then your beamformer is faulty.

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: problem with ge voluson 730 pro

June 19, 2016 02:59

If the system recognise the probe when you connect it but you do not have any image the first think you have to check is the power supply and then the BF board.