Edwin, error 1004 is related to the small drive relay board, but it could be a number of things, check the swith on the selonid for the large and small field, check your compensator movement. You can call me and can give guide you to several checks. Hope this help.
1004 - Small drives error. Check opto sensors on Collimator
Small Drives Troubleshooting
There are two types of Small Drives problems. The first involves a drive that does not move at all, and the second
involves a drive that reaches position, but the system does not recognize the fact. These errors will usually be
reported by a 2004 Error Code in the system error log. You must view the log to identify which drive is causing the
problem (Filter, Compensator, Collimator Jaws (blades), or Beam Limiter)
No Drive Motion:
1. Check 120 VAC on Collimator Relay Board (Is Neon indicator on?)
2. Check 5 VDC supply in XSC Chassis which goes to Collimator/Rotor Control Board
3. Problem will probably be in the Collimator/Rotor Control Board, Collimator Relay Board, or Collimator.
Drive moves, but system doesn’t recognize:
1. Check opto Sensor on affected drive. Might be able to swap it with the sensor from another drive.
2. Check Feedback cables to Small Drives Board
3. Problem will probably be in the Collimator/Rotor Control Board, Collimator Relay Board, or Collimator.
You can manually jumper the drives to rotate by shorting the Test Points below to RS Ground on the collimator
Rotor Control Board. Where there are two test points, one corresponds to each possible motor direction.
thank you mr. shannon; you were right, it was one of the sensors, that has a lot of dust by lack of maintenance, it was blown with air , and guala the problem desapear. thanks again
edwin
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
Edwin Rodriguez
PQ5000 error code 1004
July 12, 2008 02:49
I have a customer that has the following error code during the scout secuence, error 1004. any comment / assitance will be appreciated.
Edwin
to rate and post a comment
Ernesto Guerra
Re: PQ5000 error code 1004
July 12, 2008 07:29
Edwin, error 1004 is related to the small drive relay board, but it could be a number of things, check the swith on the selonid for the large and small field, check your compensator movement. You can call me and can give guide you to several checks. Hope this help.
to rate and post a comment
Edwin Rodriguez
Re: PQ5000 error code 1004
July 14, 2008 10:33
thank you for the info. by any chance did you worked in GE. your name is familiar to me. a guy from mexico has your name. Edwin
to rate and post a comment
Borhan Kalash
Re: PQ5000 error code 1004
July 13, 2008 11:15
Edwin,
you can do the command smadd that tests the small drives.
good luck
to rate and post a comment
Marshall Shannon
Re: PQ5000 error code 1004
July 14, 2008 02:34
1004 - Small drives error. Check opto sensors on Collimator
Small Drives Troubleshooting
There are two types of Small Drives problems. The first involves a drive that does not move at all, and the second
involves a drive that reaches position, but the system does not recognize the fact. These errors will usually be
reported by a 2004 Error Code in the system error log. You must view the log to identify which drive is causing the
problem (Filter, Compensator, Collimator Jaws (blades), or Beam Limiter)
No Drive Motion:
1. Check 120 VAC on Collimator Relay Board (Is Neon indicator on?)
2. Check 5 VDC supply in XSC Chassis which goes to Collimator/Rotor Control Board
3. Problem will probably be in the Collimator/Rotor Control Board, Collimator Relay Board, or Collimator.
Drive moves, but system doesn’t recognize:
1. Check opto Sensor on affected drive. Might be able to swap it with the sensor from another drive.
2. Check Feedback cables to Small Drives Board
3. Problem will probably be in the Collimator/Rotor Control Board, Collimator Relay Board, or Collimator.
You can manually jumper the drives to rotate by shorting the Test Points below to RS Ground on the collimator
Rotor Control Board. Where there are two test points, one corresponds to each possible motor direction.
to rate and post a comment
Edwin Rodriguez
Re: PQ5000 error code 1004
July 14, 2008 07:26
thank you mr. shannon; you were right, it was one of the sensors, that has a lot of dust by lack of maintenance, it was blown with air , and guala the problem desapear. thanks again
edwin