PQ 2000 slice count
November 10, 2011 09:54
How can I check to Gantry Rotating count and X-Ray Tube slice count?
to rate and post a comment
Borhan Kalash
Re: PQ 2000 slice count
November 13, 2011 11:00
Dear Sanchez, if your CT has XSC+ then go to regulas do: cddiag then: exposureCoun note that the C is cabital letter. Regards