Polydoros SX50 does not initialize, it stop on segment b

October 23, 2011 08:56

Hi again,

We recive another Polydoros SX50 for install, it had seen it running when uninstall on the hospital.

The changes we know are, the tube had been changed for a newer, we know we need to configure it, but we think it is going to be after the equipment start. are we right?

There for was no more changes, we turn it on this week after site installation.
In the 7segment display start counting and it stops at b (Fluoroscopy/radiography enabled), and something like a beep sounds on the generator, it seems to came from D115 Board.

From manual the next step is h.-ZB (exposure preparation) ON.
For test we put off the S2 Swich ZB from D100, then turn on again, the 7 digit display stop at b too, but without the beep sound.
We check the voltage form D115, for high voltage.

We find between X5 and X6 589 Volts DC, At X21 conector we find 7 Volts on pin 3 and 4

The led V114 Off, never turn on.

On X10, pin 1 and 3 had 29 volts AC each


Please any idea of wish is the problem, i will be greatfull.

Thanks in advance to everyone

Log inor Register

to rate and post a comment

(56)
(27)

ADRIAN GUTIERREZ

Re: Polydoros SX50 does not initialize, it stop on segment b

October 23, 2011 10:54

.( Update 23/10, 20:08 The sound came from K5 on D115, it get open and stop the supply form A1 to A2. i restart the system, at begining it had the K5 ir order, A1 to A2 with 29v in each side. When counter get b, the K5 gets open and cut supply to A2, remain 29v in A1)

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

Re: Polydoros SX50 does not initialize, it stop on segment b

October 24, 2011 10:00

Normally if stops at B the generator is ready and the error comes from the peripherals equipment

Log inor Register

to rate and post a comment

(56)
(27)

ADRIAN GUTIERREZ

Re: Polydoros SX50 does not initialize, it stop on segment b

October 24, 2011 06:30

Thanks, i find part of the error, comes from D115. But i steel cant solve it

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

Re: Polydoros SX50 does not initialize, it stop on segment b

October 25, 2011 10:00

D115 on the LX. The Anode Starter?

Log inor Register

to rate and post a comment

(56)
(27)

ADRIAN GUTIERREZ

Re: Polydoros SX50 does not initialize, it stop on segment b

October 24, 2011 09:28

Sorry everyone, is not an SX50 is an LX 30/50

Thanks

Log inor Register

to rate and post a comment

(42)
(3)

Doug Lally

Re: Polydoros SX50 does not initialize, it stop on segment b

November 29, 2011 10:19

call a guy name dave at first call parts they give support, in hopes you will need a board or something.His number is 540-520-2207 He just helped me with a similar problem. Mine was D200 board on a LX 50. You could also try reloading software.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats

Próximos eventos de leilão

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)