Tags: PACS / DR / CR / Digital

karim abdelkarim

Please help with Agfa CR35 X

March 12, 2018 09:48

We have a problem with the agfa CR 35 X . it gives the error code 20442 , and the image looks distorted with a lot of horizontal lines .
This Error occurs occasionally .

Log inor Register

to rate and post a comment

Zhang r.s

re: Please help with Agfa CR35 X

March 13, 2018 09:06

(1) Enter the Service menu

(2) Select
4 Show error
1 Explain error code

(3) Select in error code history corresponding entry
(4) Reason, meaning and cure of error are listed

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: Please help with Agfa CR35 X

March 13, 2018 09:07

20442
E_SCD_SCB_DMA_BUF_OVERFLOW
Meaning : There was at least one DMA ring buffer overflow during scan.
Reason : The image data was not stored away quick enough by the
following buffer acquisition (BAC) software.
Bad disk, system overload, ...

Log inor Register

to rate and post a comment

(13)

hasan raza

re: Please help with Agfa CR35 X

May 30, 2018 09:40

contact me my email address faisal.medical.equipments@gmail.com

Log inor Register

to rate and post a comment