Philips Epiq 7

(5)

Kekchiew Bong

Philips Epiq 7

May 31, 2021 01:41

Does any one experience of Diagnostic Code 207 after replaced both of the patient data hardisk? This code appears shortly after select presets for X5-1 or X7-2 transducer only.

Log inor Register

to rate and post a comment

(25)
(3)

heretohelp

re: Philips Epiq 7

June 02, 2021 12:53

This will most likely need a software reload. While the patient data is stored on the two patient hard drives, the patient database is stored on the OS drive. So, when the system boots up, the patient database stored on the OS drive does not match the patient data on the patient hard drives. If you are lucky, you may be able to clear this by running a patient database reset.