Philips Classic shutoff problem
January 14, 2010 10:35
I'm having an intermittent problem with the generator wanting to turn its self off when the rotor is starting. I noticed yesterday that the center volt pack was dropping during the time the rotor was starting. The unit turned back up and has been working fine sense. Has anyone had a problem similar to this before? Philips Diagnostic 71 system.
to rate and post a comment
Scott Deason
Re: Philips Classic shutoff problem
January 20, 2010 09:43
most of my problems during boost have been filament transistors.
Scott 205-290-0447
to rate and post a comment
Tom Walker
Re: Philips Classic shutoff problem
January 27, 2010 01:09
Thanks! I wasn't thinking about the problem as being an mA problem. I thouhgt that there may be a rotor problem. The center pack (Just the center one) drives down and shuts the system off. It only happens during rotor up, and it's intermittent.