Over 150 Missouri Auctions End Today - Bid Now

Philips BV 25

Tags: C-Arm

Aftab

Philips BV 25

December 28, 2012 09:57

Dears All,
I have exposure problem with BV 25 C-arm, i also doing to exposure from hand and foot switch but not produce radiation and not showing any error. Please help me what i do. i'm waiting for your reply.
Regards,
Sohail
email sohail_aftab@live.com

1 person liked this
Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Philips BV 25

December 29, 2012 01:03

They both go yo SE11 pcb, follow one or the other. Foot sw SE11 X2:13 out X2:4 to SE 33 X1:J out X1:p to SE11 X2:V. Do you have all your PS OK?

Log inor Register

to rate and post a comment

(52)
(9)

Muhammad Arsalan

re: Philips BV 25

December 29, 2012 01:03

Hi Sohail
Let me check your exposure switch is working, is there x-ray indication is showing on during exposure or not, if all this indication is ok check tube, if you need any my help let me know.
Regards,
Arsalan

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Philips BV 25

December 29, 2012 04:25

To check functionality of the switches are easy, so if there is radiation indicator lamp on the system it must be 'ON' when you press any exp. button. Also, the system generates a sound during any button pressed. The question now that if there is no radiation or there is no image. First you must check radiation output on the collimator. Electronic kV meter or DAP meter or a simple intensifying screen in a dark room is suitable to judge.
If radiation is present then the I.I and camera has the problem. I.I fault may not produce error. Fault in X-Ray system generates error message most probably except the problem is within the exposure command line.
So, separate the source of your trouble. Good luck.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

JOHNSON & JOHNSON MONARCH QUEST - For Sale $299k or Lease $6k Per Month, 48 Months

Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.

stats

Próximos eventos de leilão

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)