Felix Donoghue

Philips BV 25

July 16, 2004 12:28

I have a problem with a BV 25 which takes very high KV in auto mode after tracing out I found that the board WN 11 which is the camera control board is supposed to give an output of 1.5 volts between MP2 and 0 and you have to adjust the trimpot TAVORG to get 1.5 volts, however I found that it was giving only 0.3 volts and after adjusting TAVORG it was giving a max of 1.2volts after which it started behaving properly, however it is not getting adjusted to 1.5volts as required any reason as to why ? Thanks

Log inor Register

to rate and post a comment

(2)

Krikor Najarian

Re: Philips BV 25

July 17, 2004 02:18

Felix

I have a BV25 which is not working. I do not have a service manual to repair. (far worse then your problem :)). Can I have/buy a copy of your manual?

Regards
Krikor Najarian

krinaj@yahoo.com

Log inor Register

to rate and post a comment

(12)

Abdul Hamid Chaudhry

Re: Philips BV 25

July 27, 2004 09:49

Dear Friend,
Have you Fix the Problem of your BV-25?.
Kindly Inform ASA possible.
Best Regards to all of you

medkt@lhr.paknet.com.pk

Log inor Register

to rate and post a comment

(38)
(1)
(16)

Felix Donoghue

Re: Philips BV 25

July 28, 2004 07:49

yes problem is still unsolved.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI Solutions to Meet Your Team’s Unique Needs

From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729

stats

Próximos eventos de leilão

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (218 lots)