I am working on a Philips Br BigBore.
There will be errors in the operation of the machine. The information is:S_CTRL_LINK_DMC_GOT_CRC_FRAMING_ERROR
May appear in standby mode, Gantry start, X-ray exposure, Gantry and couch movement.
I exchanged the Stator Optical Transceiver,There is no change in the fault phenomenon.
I exchanged the Rotor Optical Transceiver,The trouble is more serious.
I want to know :1.Can the Stator Optical Transceiver and Rotor Optical Transceiver be interchanged?
2.After exchang the Rotor Optical Transceiver,must be Program the Rotor Transceiver interface box ?
Thank you very much for any suggestions!
re: Philips Brilliance BigBore CT Control Link ERROR
April 06, 2021 10:43
Hello,
Liquid bearing technology is very complicated is necessary read the newsletters from GE for understand it.
How you replace the rotor board, normally think is the next step is replace the tube (And is normal to get worry about it cuz tube is too much expensive). First condition is check if the tube mechanical are working fine, for do that you need meter the resistance of the filament and check it’s fine ; second is check the voltage signal come from the rotor board sometimes it get some leaks cause that problem. In final you need send the tube for at good factory to re check it.
If you needed more help we are a company with wide support and specialist on tubes and repair it. You can contact us.
Good Luck!
re: Philips Brilliance BigBore CT Control Link ERROR
April 12, 2021 10:00
On the big bore , The stationary and rotating can be exchanged. This is the cam bus in one direction, rotating to stationary, the other direction is not used. The transmitters go bad. They should be about the same brightness or it will not work. There is an LED on the ghost board that is for communication, rotate the gantry and watch it. Should not change, I forget if on or off. I have a bad one with 3 of the 5 channels bad on the stationary side.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
wang xin
Philips Brilliance BigBore CT Control Link ERROR
April 05, 2021 11:57
I am working on a Philips Br BigBore.
There will be errors in the operation of the machine. The information is:S_CTRL_LINK_DMC_GOT_CRC_FRAMING_ERROR
May appear in standby mode, Gantry start, X-ray exposure, Gantry and couch movement.
I exchanged the Stator Optical Transceiver,There is no change in the fault phenomenon.
I exchanged the Rotor Optical Transceiver,The trouble is more serious.
I want to know :1.Can the Stator Optical Transceiver and Rotor Optical Transceiver be interchanged?
2.After exchang the Rotor Optical Transceiver,must be Program the Rotor Transceiver interface box ?
Thank you very much for any suggestions!
to rate and post a comment
Gustavo Volpi
re: Philips Brilliance BigBore CT Control Link ERROR
April 06, 2021 10:39
1. If is exact the same, yes, you can.
2. Depends on the model, run diagnostics first and see the results.
to rate and post a comment
Gustavo Volpi
re: Philips Brilliance BigBore CT Control Link ERROR
April 06, 2021 10:43
Hello,
Liquid bearing technology is very complicated is necessary read the newsletters from GE for understand it.
How you replace the rotor board, normally think is the next step is replace the tube (And is normal to get worry about it cuz tube is too much expensive). First condition is check if the tube mechanical are working fine, for do that you need meter the resistance of the filament and check it’s fine ; second is check the voltage signal come from the rotor board sometimes it get some leaks cause that problem. In final you need send the tube for at good factory to re check it.
If you needed more help we are a company with wide support and specialist on tubes and repair it. You can contact us.
Good Luck!
to rate and post a comment
Bruce Barton
re: Philips Brilliance BigBore CT Control Link ERROR
April 12, 2021 10:00
On the big bore , The stationary and rotating can be exchanged. This is the cam bus in one direction, rotating to stationary, the other direction is not used. The transmitters go bad. They should be about the same brightness or it will not work. There is an LED on the ghost board that is for communication, rotate the gantry and watch it. Should not change, I forget if on or off. I have a bad one with 3 of the 5 channels bad on the stationary side.
to rate and post a comment