i suddenly got error X26 during scanning which went of after cleaning of HT cable plugs and tube which were dry and had some deposit on it.now , after a week it(X26) has started coming again .Is it defective tube or can be something else
Did you put new grease on the HT plugs.
we use DOW Croning #4.
If you don't do this you will get the same problem.
If you need a tube I have a Gs 1580 Eimac tube
in a B180 Housing.
From skilled GE depot repairs to patient-ready accessories, Southeastern Biomedical specializes in GE to keep your facility safe, efficient, and cost-effective. 888-310-7322.
sunil Kumar
phhilips tomoscan CX
August 19, 2004 11:16
i suddenly got error X26 during scanning which went of after cleaning of HT cable plugs and tube which were dry and had some deposit on it.now , after a week it(X26) has started coming again .Is it defective tube or can be something else
to rate and post a comment
Bill Blackford
Re: phhilips tomoscan CX
August 19, 2004 01:06
Did you put new grease on the HT plugs.
we use DOW Croning #4.
If you don't do this you will get the same problem.
If you need a tube I have a Gs 1580 Eimac tube
in a B180 Housing.
to rate and post a comment