phhilips tomoscan CX

sunil Kumar

phhilips tomoscan CX

August 19, 2004 11:16

i suddenly got error X26 during scanning which went of after cleaning of HT cable plugs and tube which were dry and had some deposit on it.now , after a week it(X26) has started coming again .Is it defective tube or can be something else

Log inor Register

to rate and post a comment

(14)

(19)

Bill Blackford

Re: phhilips tomoscan CX

August 19, 2004 01:06

Did you put new grease on the HT plugs.
we use DOW Croning #4.
If you don't do this you will get the same problem.
If you need a tube I have a Gs 1580 Eimac tube
in a B180 Housing.

Log inor Register

to rate and post a comment