OEC 8800 Problem

October 16, 2006 04:29

We are having problems with an OEC 8800, in ocasions the image of the monitor disapears, you still having X-ray but not image, and you have to reset the system to do it work again. We check the monitor and it is ok. Any help will be appreciate. Daniel Fernndez

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

Re: OEC 8800 Problem

October 23, 2006 09:58

This could be several problems, one I would check first is the II power supply, could also be the tube arced, also a cable could be broken going from the CCD to the video processor bd, then another check would be to reformat your image drive you might have bad sectors.

Log inor Register

to rate and post a comment

(46)
(6)

Daniel Fernandez

Re: OEC 8800 Problem

October 30, 2006 03:38

Thank you Lenny.
We have checked any of this parts, but not others.

Log inor Register