Hi all,
Wich is the procedure for replacing a defective Memoskop CX?
In the USB port is a hardware key. Customer license CD is on site. What to do next?. The defective Memoskop shows horizontal stripes randomly on the image at startup and when I release radiation. Even without connecting the video input jack, the same stripes appear.
Thank you in advance!
Your troubleshooting appears to have eliminated the TV camera (removal of the Video-In), so the next step is to replace the Memoskop CX. Do you have p/n 7142537 ? If you buy a replacement from us, you would transfer your USB license dongle to the replacement Memoskop, then use your License CD to transfer your License to the replacement Memoskop. I can email you the procedure (what is your email?) Be sure toi return your defective Memoskop CX to get your exchange credit.
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Adrian Mihaita
Memoskop CX replacement
November 25, 2019 05:59
Hi all,
Wich is the procedure for replacing a defective Memoskop CX?
In the USB port is a hardware key. Customer license CD is on site. What to do next?. The defective Memoskop shows horizontal stripes randomly on the image at startup and when I release radiation. Even without connecting the video input jack, the same stripes appear.
Thank you in advance!
to rate and post a comment
dms
re: Memoskop CX replacement
November 26, 2019 12:32
which Siemens carm model ???
to rate and post a comment
Adrian Mihaita
re: re: Memoskop CX replacement
November 29, 2019 01:58
Siremobil Compact L sn 31087/2012
Memoskop CX/98202907-CX/sn S79508
to rate and post a comment
Wayne Horsman
re: Memoskop CX replacement
November 27, 2019 09:59
Your troubleshooting appears to have eliminated the TV camera (removal of the Video-In), so the next step is to replace the Memoskop CX. Do you have p/n 7142537 ? If you buy a replacement from us, you would transfer your USB license dongle to the replacement Memoskop, then use your License CD to transfer your License to the replacement Memoskop. I can email you the procedure (what is your email?) Be sure toi return your defective Memoskop CX to get your exchange credit.
to rate and post a comment