Hello everyone, I have a DUO systems, run QA and all tests PASS, but ligth sent RED. Reviewing QA trend graphics Low ma ratio is low.
It is possible move the lower scale of that graph?
What does represent that value?
Any tip to test?
I will glad if someone could help me on it.
You may have tube or detector that is on its way out (if it hasn't failed completely already). I would start by resetting the QA Means and attempt another QA.
I would also suggest emailing us a QA printout and the QA trend graphs to info@dexasolutions.com. Once we have that we can provide some further advice.
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
villa
Low ma ratio
December 30, 2018 02:41
Hello everyone, I have a DUO systems, run QA and all tests PASS, but ligth sent RED. Reviewing QA trend graphics Low ma ratio is low.
It is possible move the lower scale of that graph?
What does represent that value?
Any tip to test?
I will glad if someone could help me on it.
Thank you in advance
to rate and post a comment
Ben Williams
re: Low ma ratio
January 01, 2019 08:32
Hi Villa,
You may have tube or detector that is on its way out (if it hasn't failed completely already). I would start by resetting the QA Means and attempt another QA.
I would also suggest emailing us a QA printout and the QA trend graphs to info@dexasolutions.com. Once we have that we can provide some further advice.
Thanks,
Ben
to rate and post a comment