Lorad Multicare platinum stereo unit error 27

(38)
(6)

John Mendoza

Lorad Multicare platinum stereo unit error 27

May 13, 2008 05:28

I get an error 27 intermittently. It can happen once a week or once a month. I have changed out the filament protection board and calibrated the filaments many times. If you have any information to help me out I would greatly appreciate it.
Thanks,
John
j.mendoza@scmes.com

Log inor Register

to rate and post a comment

(110)
(10)
(3)

Deon Armes

Re: Lorad Multicare platinum stereo unit error 27

May 20, 2008 01:16

John,
This is an excert from the service manual.
27- CONTROL- Filament check in boost routine not within 12.5% of adaptive
table of DAC values. Good luck!
Deon

Log inor Register

to rate and post a comment

(40)
(15)

John Edwards

re: Re: Lorad Multicare platinum stereo unit error 27

July 19, 2014 09:25

Do you have the service login and password for this unit?

Log inor Register

to rate and post a comment