Lorad Multicare platinum stereo unit error 27
May 13, 2008 05:28
I get an error 27 intermittently. It can happen once a week or once a month. I have changed out the filament protection board and calibrated the filaments many times. If you have any information to help me out I would greatly appreciate it.
Thanks,
John
j.mendoza@scmes.com
to rate and post a comment
Deon Armes
Re: Lorad Multicare platinum stereo unit error 27
May 20, 2008 01:16
John,
This is an excert from the service manual.
27- CONTROL- Filament check in boost routine not within 12.5% of adaptive
table of DAC values. Good luck!
Deon
to rate and post a comment
John Edwards
re: Re: Lorad Multicare platinum stereo unit error 27
July 19, 2014 09:25
Do you have the service login and password for this unit?