Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

LORAD MIV power problem

(99)
(11)
(6)

Yojin Ramones

LORAD MIV power problem

April 04, 2009 02:21

I am work with a problem with a MIV it is not power on.

check all fuse all is ok . in the power card the led is green and Tp1 have 24 volts but tp2 no have.

check all security swith and all is ok.

I suspet problem in power card?? I could apreciated clue thanks firends

Log inor Register

to rate and post a comment

(259)
(59)
(3)

Jose Vallejos

Re: LORAD MIV power problem

April 08, 2009 02:54

Hi, i have a service manual for this machine, please contact me, javv2005@hotmail.com
Best Regards

Log inor Register

to rate and post a comment

preecha chaipet

Re: LORAD MIV power problem

April 09, 2009 09:52

i have too, can you check switching power supply

Log inor Register

to rate and post a comment

(99)
(11)
(6)

Yojin Ramones

Re: LORAD MIV power problem

April 11, 2009 05:06

Thanks jose

The led green is on in the power card and test this card and it is ok.

power switch in the console or brake in the gantry?

regards

yojin


Log inor Register

to rate and post a comment

(16)
(5)
(1)

Hector Morales

Re: LORAD MIV power problem

August 03, 2009 05:43

Hi, do you have a copy of service manual for Lorad MIV? Can you give it by email?
hmoralespsm@gmail.com

Thanks and best regards

Log inor Register

to rate and post a comment

(99)
(11)
(6)

Yojin Ramones

Re: LORAD MIV power problem

April 17, 2009 01:35

Thanks friend for the suppor
a problem was a limit swtch of colum that it was active.

I release and the system work again


thanks to all

yojin

Log inor Register

to rate and post a comment

antoninus andres

re: Re: LORAD MIV power problem

November 10, 2014 04:34

Hi sir
Where is the limit switch located in the machine? U have same problem, hope we you can help me..

Log inor Register

to rate and post a comment

(99)
(11)
(6)

Yojin Ramones

re: re: Re: LORAD MIV power problem

November 11, 2014 10:17

Try move the column manually for release of the switch limit position. And check if it work again

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Próximos eventos de leilão

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)