Manuel Casanova

LORAD M3 ERROR SYSTEM

May 19, 2011 11:09

HELLOW:

I HAVE A ERROR CODE FOR LORAD M3: "SYSTEM ERROR", I EXCHANGE THE CPU BOARD, BUT THE PROBLEM PERSIST.

DO YOU HAVE ANY IDEAS??

BEST REGARDS

Log inor Register

to rate and post a comment

(27)
(3)
(5)

James McGinty

Re: LORAD M3 ERROR SYSTEM

May 20, 2011 03:05

Take a look at the low voltage power supply..

J. McGinty

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: LORAD M3 ERROR SYSTEM

May 22, 2011 09:09

Thank´s, I´m fixed the problem at LPS.

Thank´s a lot for everyone

Log inor Register

to rate and post a comment

(26)
(5)

Jose Garcia

Re: LORAD M3 ERROR SYSTEM

July 18, 2011 12:02

Can you provide service manual for Lorad MIII

thanks

1 person liked this
Log inor Register

to rate and post a comment

preecha chaipet

Re: LORAD M3 ERROR SYSTEM

September 11, 2011 07:24

hi friend
you have problem with microprocessor board .watchdog timeout you can fix problem by repair one or more this components on board
pls let me know on r_technologist@yahoo.com
good luck

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Próximos eventos de leilão

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)