Over 800 Total Lots Up For Auction at Four Locations - VA 07/21, CA 07/22, KS 07/24, CA 07/31

Image quality-SR7000

sunil Kumar

Image quality-SR7000

October 07, 2005 09:22

I am getting darker images with window shift towards lower valuesiewater phantom images showing CT values at -55 at 120 KV ,2s and 4 sec scan.the images get pretty ok at 1sec.ct value -25.software,ct number calib has no effect.KV and ma waveform are o.k.Any suggestions

Log inor Register

to rate and post a comment

Ralph Childs

Re: Image quality-SR7000

October 08, 2005 12:59

Sunil,

Are the low CT numbers throughout the phantom or mainly in the top region??? If they are in the top, you need to check phantom vertical position when doing the calibration.

If they are consistent throughout the phantom then you have a number or things to check:

1) How old is the tube? Is the dose output dropping off due to tungsten splatter on the port?
2) What is the condition of the collimator filter? Does it have cracks in the surface?
3) Was this phantom okay before? Make certain that the "filler" is actually dsitilled water and not alcohol.
4) You need to verify your actual rotation speeds. It alomost sounds like your slower speeds are incorrect.

Let me know if I can help you.

Scooter
scanner1@eim.ae

Log inor Register

to rate and post a comment

(86)
(5)

SHIJU KARAKKOOTTIL

Re: Image quality-SR7000

October 08, 2005 04:13

Hi
Conform your tube centering.
Regards,
Shiju.
shiju@cura.co.in

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats
advertisement

Próximos eventos de leilão

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (218 lots)