fluor is ok,only no image display on monitor.
No ERROR display.open the cover of I.I,led V11 light,some time V8 light.Generated from the external 11.3 V. Generated from the external 5.5 V. are ok.use S2 3-6 off to display the test image=>ok.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Sin Man Lau
Iconos R100 problem.
October 16, 2009 05:38
fluor is ok,only no image display on monitor.
No ERROR display.open the cover of I.I,led V11 light,some time V8 light.Generated from the external 11.3 V. Generated from the external 5.5 V. are ok.use S2 3-6 off to display the test image=>ok.
change VIDEOMED DIC .