Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

I have problem with picker eclipse 1.5t

Selcuk Akin

I have problem with picker eclipse 1.5t

March 21, 2007 05:42

I have problem with picker eclipse 1.5T mr.
The machine works with channel 1 coil (head coil,volume neck coil, no problem)but it doesnt work with channel 4 coil (flex body coil,ctl coil,atlas head neck coil. doesnt work)
what problem is this?
Has anyone exeperence with this kind of problem.
Thanks.

Log inor Register

to rate and post a comment

(317)
(22)
(8)

Joseph Jenkins

Re: I have problem with picker eclipse 1.5t

March 21, 2007 12:05

Contact Dave adams @ dave3dt@yahoo.com

Tell him jay sent you

Log inor Register

to rate and post a comment

Selcuk Akin

Re: I have problem with picker eclipse 1.5t

March 26, 2007 06:33

thanks for your help

Log inor Register

to rate and post a comment

(93)
(7)

Edward Yan

Re: I have problem with picker eclipse 1.5t

March 22, 2007 08:16

Selcuk Akin,

You have a faulty Receiver board in con/recon cabinet. There are two Receiver boards. If you swapped them, the channel 4 coils should work. Sometimes when you reseating the board properly, problem might disappear.

Edward Yan, B.S.E.E.

Log inor Register

to rate and post a comment

Selcuk Akin

Re: I have problem with picker eclipse 1.5t

March 26, 2007 06:41

thanks for you help

Log inor Register