Dean Kenney

HSA rp1 2 second scan problem

January 18, 2006 04:06

okay a year and a half another company sold and installed a rp1 with this problem and i stumbled across it after taking over the service contract. In "S" tilt only top towards rear of gantry bottom towards table the unit will scan a patient in every speed but 2 second with a annode overcurrent error in all other respects the unit runs great. Also any degree from 2 to whatever if its "S" it errors. 1 sec helical all works in "S" tilt Hav'nt spent any toubleshooting time just checked the error because it was the end of a long trip and the customer did a by the way an hour before the plan left on of course friday thank guys

Log inor Register

to rate and post a comment

(47)

Mike Lynch

Re: HSA rp1 2 second scan problem

January 18, 2006 10:36

You probably have an intermittent power supply problem or STC board problem. Try reseating the board and check its power supply.