you need to check the alignment of the table and system. make sure the system is level and not tilted. also check the image to see where is the problem, you may see some part of the metal rim of the table in the picture.
Dominador Cagulada
Hologic QDR Discovery
January 23, 2019 09:56
System Test Fail during QC. Radiographic Uniformity acceptable = below 2 SD. Please need help
to rate and post a comment
MRAdmin
re: Hologic QDR Discovery
January 26, 2019 09:39
you need to check the alignment of the table and system. make sure the system is level and not tilted. also check the image to see where is the problem, you may see some part of the metal rim of the table in the picture.
to rate and post a comment
Dominador Cagulada
re: re: Hologic QDR Discovery
February 06, 2019 09:09
Hi good day. System is level, i check image for metal and found nothing. Sometimes system pass but qc failed due to bmc & area are too high.
to rate and post a comment
GNES
re: re: re: Hologic QDR Discovery
February 07, 2019 09:04
do you did the planning od detector, can you inside of service fro calibrate
to rate and post a comment
GNES
re: re: re: Hologic QDR Discovery
February 07, 2019 09:05
write me gnesker@gmail.com
to rate and post a comment
MRAdmin
re: re: re: Hologic QDR Discovery
February 07, 2019 09:05
your system need full calibration and x-ray alignment check
to rate and post a comment
GNES
re: re: re: re: Hologic QDR Discovery
February 07, 2019 02:18
hello mradmin if you need some help i work with hologic equiment dimension, selenia , multicare affirm, afirm prone
to rate and post a comment