Head Coil Problem - tr status error 04
March 26, 2007 06:55
I have problem with the head coil of a picker 1.5 t mr.
The error appears on mr screen 'tr status error 04'
What is this?
Thanks
to rate and post a comment
Bruce Smith
Re: Head Coil Problem - tr status error 04
March 26, 2007 07:31
The most likely cause is a Pin Switch failure.
to rate and post a comment
Richard McCrocklin
Re: Head Coil Problem - tr status error 04
March 27, 2007 08:58
I found it to gome from either of 2 places.
T/R pcb which mixes the incoming sigs from coil, or the connectors. They get dirty and worn out. I had to replace a individual pin on the bed that the head coil uses. The coil is only using 3 pins (quad), single coils only use 2 pins (pwr and sig)