Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

HDI 5000 error message 0017

Tags: Ultrasound

(8)

Issouf NIKIEMA

HDI 5000 error message 0017

June 15, 2018 05:13

Hi Team
Need Help to diagnose HDI 5000 ultrasound.
This is the error message appear :" an error has been detected. pease record the following information and call ATL technician support"
"Diagnatic Number: 0017"
Also I need the Service manual.
Email: nikiema3105@gmail.com

Log inor Register

to rate and post a comment

(222)
(74)
(5)

Marko Shevchuk

re: HDI 5000 error message 0017

June 17, 2018 11:27

Hi Issouf,
0017 bUtilStatus.c "Slot configuration error on boot up.
Review the previous error messages and the suspected slot and board will be identified. Either the configuration was read incorrectly which would be a board failure or the board was not seated correctly. Another possibility is that the data files that describe how the center plane is configured (brdList.xml, ctrP96.xml, sltP96.xml) are invalid or corrupt.
System reported upper and lower UIF’s not recognized. FSE swapped the PSM, Lower UIF, and the center plane. The center plane was found defective.

Log inor Register

to rate and post a comment

(8)

Issouf NIKIEMA

re: re: HDI 5000 error message 0017

June 18, 2018 10:26

Hi,
thank you for response. the first error was 0002 before he show 0017. physically cards are ok,we cleaned and replace correctly but this don't solve problem.
can you help with service manual. thanks

Log inor Register

to rate and post a comment

(222)
(74)
(5)

Marko Shevchuk

re: re: re: HDI 5000 error message 0017

June 18, 2018 04:32

Hi Issouf,
1. 0002 Error
bCommonFormat.c - FEC alert message - the front
end has stopped scanning because of some power monitor
problem. This could be a hardware problem with the
scanhead, the channel boards, the AIM, or the power
supplies, but this could also be a system software problem.
Please review FEC errors just prior to the event for
additional hints or board ID's.

2. You can get more information if you connect the laptop
to the second connector from the top of the CPU PCB using serial cable.
If you will have any question please write me to markoshv@ukr.net

Log inor Register

to rate and post a comment

(8)

Issouf NIKIEMA

re: re: re: re: HDI 5000 error message 0017

June 19, 2018 08:04

OK, thank you

Log inor Register

to rate and post a comment

(17)
(2)

Daouda Thiam

re: HDI 5000 error message 0017

June 17, 2018 11:27

Hi ISSOUF.
I have the reference manual, if it's good for you, I'll send you

Log inor Register

to rate and post a comment

(8)

Issouf NIKIEMA

re: re: HDI 5000 error message 0017

June 18, 2018 10:22

Hi,
It's ok for me please send it

Log inor Register

to rate and post a comment

(17)
(2)

Daouda Thiam

re: re: re: HDI 5000 error message 0017

June 18, 2018 12:03

please check your email

1 person liked this
Log inor Register

to rate and post a comment

(8)

Issouf NIKIEMA

re: re: re: re: HDI 5000 error message 0017

June 19, 2018 08:04

Hi,
I recieved the ref manual. Thank you

Log inor Register

to rate and post a comment

stats
DOTmed text ad

CC Medical : Ultrasound systems You Can Trust

We work with hospitals providing reliable ultrasound systems, probes & parts. By offering sales, service, rentals & repairs, we help our clients access high quality imaging systems while keeping costs low & support simple. Please call 760-834-4777!

stats
advertisement

Próximos eventos de leilão

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)