I have a minor problem with a GE proteus generator after taking an exposure and then removing the cassette from the table bucky the system intermittently plays a tune a series of notes? . Has anyone come across this before . It doesn't seem to affect system operation.
Regards
Mike
to rate and post a comment
xrayrepairdude
re: Ge proteus
November 27, 2015 08:12
That sounds like the bucky drive motor that has a delay on driving the grid back to home position after the exposure. It may eventually fail and the generator won't expose on that bucky.
Ken
to rate and post a comment
Michael Kakoulli
re: re: Ge proteus
November 27, 2015 10:57
Hi ken,
will check it out.
Mike
to rate and post a comment
Lenny Place
re: Ge proteus
November 27, 2015 03:46
check the power supplies on the system I/F board in the top of the Jedi cabinet if the the +5 VDC is at or close to 4.85VDC in the base of there is a power supply if you look at the wires coming off the supply you will see a connector with power off reseat this connector(if you have contact cleaner spray or apply to both sides) then retest the supplies.
the intermittent tone can be a reset tone due to these supplies