GE Prodigy problem

(38)
(6)

John Mendoza

GE Prodigy problem

July 29, 2009 02:43

Hello Everyone, I have a GE Prodigy with an intermittent error- "scanner Reset". I have check the log but ther is very little information about this error. If you have any feed back it will be greatly appreciated.
I'm also looking for the service password to the latest prodigy software Ver. 11.20.068
Thanks
John Mendoza

Log inor Register

to rate and post a comment

(17)

(1)

Jailyn Primus

Re: GE Prodigy problem

September 10, 2009 08:55

Most common problem with the Prodigy is the cables. Check the bulkheads for connectivity, and the voltages to the detector. Bad shorts cause missing voltages in the upper cable bundle.

Log inor Register

to rate and post a comment

(17)

(1)

Jailyn Primus

Re: GE Prodigy problem

September 10, 2009 08:55

Short in the bulkheads or short in the wiring going to the detector. Or short in the3 upper cable bundle