Over 150 Missouri Auctions End Today - Bid Now

GE MPPU Error 84A3

(34)
(5)
(1)

Dinu Papoiu

GE MPPU Error 84A3

November 29, 2010 10:55

Hi:

On the diagnostic mode for GE MPPU generator, I get error 84A3 on the power unit. Does anyone know what this error code is?
Thank you,
Dinu

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE MPPU Error 84A3

December 03, 2010 03:17

Mppu Control to Mppu power unit communications check. Checks that an exposure can be initiated from the ECM (Advantx). Verifies feedback from the high voltage bleeder. Check comm cables and all connections to/from the Advantx, between the control and power cabinets, to the high voltage transformer.

Log inor Register

to rate and post a comment

(34)
(5)
(1)

Dinu Papoiu

Re: GE MPPU Error 84A3

December 05, 2010 09:38

Thank you very much.
I really appreciate your help.It clarified a lot of questions which I had about the exposure problem.
Dinu

Log inor Register

to rate and post a comment

stats
DOTmed text ad

CC Medical : Ultrasound systems You Can Trust

We work with hospitals providing reliable ultrasound systems, probes & parts. By offering sales, service, rentals & repairs, we help our clients access high quality imaging systems while keeping costs low & support simple. Please call 760-834-4777!

stats

Próximos eventos de leilão

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)