Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

GE Hispeed CT Error Code: 40 - 0107H Rotation Error

Tags: CT, GE

(3)

Asad

GE Hispeed CT Error Code: 40 - 0107H Rotation Error

September 21, 2020 10:37

Hello, GE hispeed NXi pro CT Scan machine When We try to calibration or Tube warmup or any scan we got the following errors:

XG Error: Phase is 5 code is 60
Detail Error Code 60-0316H
Exposure Error: Spit Max Error

we got the following error:

XG Error, Error: Phase is 2 code is 40
Detail Error Code 40-0107H
Exposure Error: Rotation error: Rotation Inverter permanent overcurrent

I will appreciate any help that could be provided.
Thank you all.
Asadullah Asad

Log inor Register

to rate and post a comment

(172)
(40)
(3)

Dmitry Dederer

re: GE Hispeed CT Error Code: 40 - 0107H Rotation Error

September 22, 2020 10:32

60-0316H
Spit Max error.
kV control has detected the
number of tube spits during
exposure has reached the
limit (see theory of
operation, software
section)
reasonably x-ray tube -Try again at various kV/mA to
confine problem.
-Troubleshoot HV section. (x-ray
tube, cable, HV tank)
-Diagnose based on tube history.

40-0107H
Rotation Inverter
permanent overcurrent
An overcurrent has been
detected and 3 restart
have been tried
unsuccessfully within a
single rotation state
1. Rotation board
2. Tube stator winding in
short circuit -> cable
(NP) or HEMIT (only
for NP++)
3. Wiring incorrect
(shorted)
-Check wiring from rotation board
to tube.
-Troubleshoot tube windings
-Replace Rotation board

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: GE Hispeed CT Error Code: 40 - 0107H Rotation Error

September 22, 2020 10:32

ERROR 60-0316H:
Description:kV control has detected the number of tube spits during exposure has reached the limit.
Cause: Faulty X-ray Tube

ERROR 40-0107H:
Description: An overcurrent has been detected and 3 restart have been tried unsuccessfully within a single rotation state.
Cause: Faulty Rotation Board, Tube Stator Winding short circuit.

RECOMMENDATION:
- Check cabling from rotation board to tube
- perform a NO LOAD test to isolate the cause
- troubleshoot tube windings
- Check anode HV cable

It looks like you have a shorted tube windings....and possibly a bad rotation board.

I would recommend replacement of X-ray tube and Rotation Board

craigwambura@outlook.com

Log inor Register

to rate and post a comment

(102)
(15)
(1)

Maurice Dilick

re: re: GE Hispeed CT Error Code: 40 - 0107H Rotation Error

September 25, 2020 11:10

To check the Stator, measure the resistance. Normal problem of probably the cable.

1 person liked this
Log inor Register

to rate and post a comment