Over 300 Virginia Auctions End Today - Bid Now
Over 300 Total Lots Up For Auction at Three Locations - Over 100 California Auctions End Tomorrow 07/22 - Bid Now, KS 07/24, CA 07/31

GE DMR Mamo E21 Heating error

ertan yagci

GE DMR Mamo E21 Heating error

November 08, 2008 07:41

I've E21 Heating Error on GE DMR Mamo Unit.
I can exposure manually but If I select any kind of auto ma or kv system cant exposure because of this error.The filament seem to me OK because of I can exposure manully .
Does anybody give us any suggestion about this error.

Log inor Register

to rate and post a comment

Ralph Childs

Re: GE DMR Mamo E21 Heating error

November 09, 2008 12:02

Ertan,

E21 is a heater software error. This is why you can expose manually without the error. It is most probably due to a large difference in the stored calibration data. If this is a used system that you are installing at a new location there is probably a large variance in the line voltage and/or frequency.

Scooter
scanner1@eim.ae

1 person liked this
Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: GE DMR Mamo E21 Heating error

November 09, 2008 01:18

Ertan:

The problem is in your AEC exposure control, check the fuse for HV for AEC, if you need the services manualls, please contact me asap.

Manuel

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE DMR Mamo E21 Heating error

November 10, 2008 04:07

It helps to know what the extended error code is, when the E21 comes up on the far right of the screen there is a six digit code XXX/XXX that can be very specific about what or where the problem is. What is the extended error code??

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE Monitor Service You Can Trust

Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.

DOTmed Certified stats

Próximos eventos de leilão

C2 Management
TIMED Ending Soon
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (229 lots)