Gufaw Hagos
GE brightspeed CT 16 slice
August 10, 2017 08:06
Dear all,
I have faced a problem with table,which in able to move in up/in position(elevation is deactivated).
how can i solve this one.
to rate and post a comment
Gabriel Cool
re: GE brightspeed CT 16 slice
August 10, 2017 11:10
Do you mean you cant move table up/down and in/out?
to rate and post a comment
Gufaw Hagos
re: re: GE brightspeed CT 16 slice
August 13, 2017 01:47
yes,to up/down
to rate and post a comment
Gabriel Cool
re: re: re: GE brightspeed CT 16 slice
August 14, 2017 10:25
Ok. Try to make table sinhronization. You need adjustment manual. Test 9 and 10. Give me your email i will send you this manual
to rate and post a comment
Gufaw Hagos
re: re: re: re: GE brightspeed CT 16 slice
August 15, 2017 10:33
may mail is gufaw.hagos@gmail.com
to rate and post a comment
Artyom Kizhakin
re: re: re: re: GE brightspeed CT 16 slice
December 26, 2017 08:42
Hi! can you send manual to dcrub@yandex.ru, we have the same brightspeed and sometimeis it's won't move the table
to rate and post a comment
CRAIG WAMBURA
re: GE brightspeed CT 16 slice
August 13, 2017 01:48
Check your MTCB and servo drives.
Most probably you have a faulty MTCB. If MTCB is ok, first check fuse F01, Then check the two servo drives.
craig@telemed-technologies.com
to rate and post a comment
Martin Martinez
re: GE brightspeed CT 16 slice
August 13, 2017 02:07
Gufaw, Hello.
Please share me the error log to try to help you.
Regards.
to rate and post a comment
Gufaw Hagos
re: re: GE brightspeed CT 16 slice
August 15, 2017 10:33
here the error log is 265304
to rate and post a comment