Error M 384 on Philips Veradius

April 04, 2018 09:11

Hi at all.
I have a problem on a Philips Veradius system.
after power on display C arm appair error M 384 end of life battery.
someone can help me ?
I need also service manual !
Thanks in advance for your answer !!

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: Error M 384 on Philips Veradius

April 06, 2018 10:01

Dear Vincent,

The Err 384 states that the built-in ESU power storage unit is defective. This unit provides the necessary power for the X-ray generator with 15 KW power. From a normal power outlet comes max. in Europe 2.3 KW.

The ESU consists of 26 individual batteries and can be ordered with the Philips spare part number 4598 009 0822.

Important, if an old charger is used this should also be replaced.

After replacement, the service software BV-Scope must be used to re-register the battery in the system. Only then does the device work correctly again.

Many greetings
Willi

Incidentally, if there is a problem with a Philips BV system, please always include the 6NC number.

This starts with Philips BV system always with 718 ??? and only with this number is a device technically identifiable and it can be worked with the proper documentation.

The software version is also important!

Log inor Register

to rate and post a comment

(5)

Vincent58

re: re: Error M 384 on Philips Veradius

April 10, 2018 08:50

Thanks for your answer. I will search, now, the service software BV-Scope 😞😞

Log inor Register

to rate and post a comment

stats
DOTmed text ad

JOHNSON & JOHNSON MONARCH QUEST - For Sale $299k or Lease $6k Per Month, 48 Months

Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.

stats

Próximos eventos de leilão

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)