Error in Checkup Somatom Esprit

Tags: CT, SIEMENS

(34)
(6)

Kian Chaves

Error in Checkup Somatom Esprit

December 19, 2017 09:16

Hello Guys,
I'm serving a client who has an older Somatom with IRS 5 and ICS 6. The client made a call because the Workstation didn't go up the Syngo. It has been reported that the CT isn't switched on for 7 months. Analyzing the IRS and ICS I found that they were running out of battery and losing the date / time. After changing the battery and setting the date and time the system has turned on perfectly. However at the time of checkup I got the following message:
Initilization failed.
Filament adaptation aborted by error.
And by viewing the Event Log the following message was reported
Invalid / inconsistent value: Cooling duration is negative! Probaly the systemtime has been changed or tubehistory is corrupt.
May someone help me?

1 person liked this
Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: Error in Checkup Somatom Esprit

December 19, 2017 03:56

Hm i think you need find tube history file on C disc. And delite all from this folder. Then reboot system

1 person liked this
Log inor Register

to rate and post a comment

(34)
(6)

Kian Chaves

re: re: Error in Checkup Somatom Esprit

March 06, 2018 02:17


I'll look for the tubehistory file and try this possible solution. Thank you very much for the information I will be starting the service now

Log inor Register

to rate and post a comment

(94)
(22)

Mario E Sarduy La Paz

re: Error in Checkup Somatom Esprit

December 20, 2017 09:15

Hi.
Maybe doing a restore the CT setting could be other solution too.
Best Regards and let us know the final result or share the eventlog file for do a better analysis of the problem.

1 person liked this
Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: re: Error in Checkup Somatom Esprit

December 20, 2017 03:37

Restore setting restored only system setting, but in this situation system uoad information from tube and thats why he has a error. For test, change date on your sistem which was before replased battery

1 person liked this
Log inor Register

to rate and post a comment

(34)
(6)

Kian Chaves

re: re: re: Error in Checkup Somatom Esprit

March 06, 2018 02:17


I'm gonna try now. I'll give you a return as soon as possible, thank you very much for the help

Log inor Register

to rate and post a comment

(34)
(6)

Kian Chaves

re: re: re: Error in Checkup Somatom Esprit

March 09, 2018 03:53

The IRS motherboard had a problem with a capacitor near the battery and CMOS, I made the switch and it's OK. As for the tubefile file I have already found it however I don't have access to delete it. What should I do to gain access?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Brand new OEM Philips & Siemens Parts at unbeatable prices

Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty! Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Próximos eventos de leilão

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (197 lots)